Feasibility study of…
Feasibility study of a scintillation sheet-based detector fo…
Purpose: This study evaluated the properties of a scintillation sheet-based dosimetry system for beam monitoring with high spatial resolution, including the effects of this system on the treatment beam. The dosimetric characteristics and feasibility of this system for clinical use were also evaluated. 
Methods: The effects of the dosimetry system on the beam were evaluated by measuring the percentage depth doses, dose profiles, and transmission factors. Fifteen treatment plans were created, and the influence of the dosimetry system on these clinical treatment plans was evaluated. The performance of the system was assessed by determining signal linearity, dose rate dependence, and reproducibility. The feasibility of the system for clinical use was evaluated by comparing intensity distributions with reference intensity distributions verified by quality assurance. 
Results: The spatial resolution of the dosimetry system was found to be 0.43 mm/pixel when projected to the isocenter plane. The dosimetry system attenuated the intensity of 6 MV beams by about 1.1%, without affecting the percentage depth doses and dose profiles. The response of the dosimetry system was linear, independent of the dose rate used in the clinic, and reproducible. Comparison of intensity distributions of evaluation treatment fields with reference intensity distributions showed that the 1%/1 mm average gamma passing rate was 99.6%. 
Conclusions: The dosimetry system did not significantly alter the beam characteristics, indicating that the system could be implemented by using only a transmission factor. The dosimetry system is clinically suitable for monitoring treatment beam delivery with higher spatial resolution than other transmission detectors.

Jaehyeon Seo, Hyunho Lee, Myonggeun Yoon*
Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Jaehyeon Seo
Environmental Radioactivity Assessment Team, Korea Atomic Energy Research Institute, Daejeon, Republic of Korea

Hyunho Lee, Sung Hwan Ahn*
Department of Radiation Oncology, Samsung Medical Center, Seoul, Republic of Korea

Myonggeun Yoon*
FieldCure Ltd, Seoul, Republic of Korea 
Study of a plastic s…
Study of a plastic scintillating plate‑based quality assu…
Introduction: The purpose of this study was to evaluate a plastic scintillating plate‑based beam monitoring system to perform quality assurance (QA) measurements in pencil beam scanning proton beam. 
Methods: Single spots and scanned fields were measured with the high‑resolution dosimetry system, consisting of a plastic scintillation plate coupled to a camera in a dark box at the isocenter. The measurements were taken at 110–190 MeV beam energies with 30° gantry angle intervals at each energy. Spot positions were determined using the plastic scintillating plate‑based dosimetry system at the isocenter for 70–230 MeV beam energies with 30° gantry angle intervals. The effect of gantry angle on dose distribution was also assessed by determining the scanning pattern for daily QA and 25 fields treated with intensity‑modulated proton therapy. 
Results: Spot size, field flatness, and field symmetry of plastic scintillating plate‑based dosimetry system were consistent with EBT3 at all investigated energies and angles. In all investigated energies and angles, the spot size measured was ±10% of the average size of each energy, the spot position measured was within ±2 mm, field flatness was within ±2%, and field symmetry was within ±1%. The mean gamma passing rates with the 3%/3 mm gamma criterion of the scanning pattern and 25 fields were 99.2% and 99.8%, respectively. 
Conclusions: This system can be effective for QA determinations of spot size, spot position, field flatness, and field symmetry over 360° of gantry rotation in a time‑ and cost‑effective manner, with spatial resolution comparable to that of EBT3 film.

Jaehyeon Seo, Myonggeun Yoon*
Department of Bio-Convergence Engineering, Korea University

Kwangzoo Chung, Youngyih Han, Sung Hwan Ahn*
Department of Radiation Oncology, Samsung Medical Center

Youngyih Han
Department of Radiation Oncology, Samsung Medical Center, Sungkyunkwan University School of Medicine, Seoul

Seonghoon Jeong
Department of Proton Therapy Center, National Cancer Center, Goyang

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University

Geon Oh, Yongha Gi, Heehun Sung
Department of BioMedical Engineering, Korea University, Seoul, Republic of Korea

Myonggeun Yoon*
FieldCure Ltd, Seoul, Republic of Korea
Correlation between …
Correlation between impulse magnitude and inhibition of cell…
This study was designed to investigate the correlation between the impulse by dielectrophoretic force applied inside a dividing cell during alternating electric fields therapy and the inhibition of cell proliferation. Distributions of the electric field and dielectrophoretic force inside a dividing cell were calculated using the finite element method of COMSOL Multiphysics. Based on the results, the average magnitude of the impulse by the dielectrophoretic force applied to the cleavage furrow inside a dividing cell placed in various directions was calculated as a function of electric field intensity at an extracellular reference point. The simulation results showed that the average magnitude of the impulse to the cleavage furrow inside a dividing cell ranged from 1.51 × 10−9 to 1.49 × 10−7 N s when tumor treating fields with an intensity ranging from 0.1 to 1 V/cm is applied at an extracellular reference point for 6 h. To verify the relationships between the impulse by the dielectrophoretic force and the inhibition of cell proliferation, the survival fractions of the four cancer cell lines were determined as a function of intensity and time duration of the electric field. The correlation between the magnitude and application time of the electric field and the survival fractions of the four cell lines showed similar trends in vitro. These results suggest that both the dielectrophoretic force and the time required for the force to act are proportionally related to the inhibitory effect on dividing cells, enabling this impulse to be used as a reference to quantify the inhibition of cell proliferation.

Geon Oh, Yongha Gi, Jinyoung Hong
Department of Bioengineering, Korea University, Seoul, Republic of Korea

Geon Oh, Boram Lee*
Department of Radiation Oncology, Inha University School of Medicine, Incheon, Republic of Korea

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University, Seoul, Republic of Korea 

Jonghyun Kim, Myonggeun Yoon*
FieldCure Ltd., Seoul, Republic of Korea 
Application of Fast …
Application of Fast Non-Local Denoising Approach in Digital …
Introduction: Chest X-ray imaging has become the most commonly used, as it is the primary method for lung cancer screening during medical check-ups. The radiation dose should be minimized to ensure that the patients are not overexposed to radiation. However, radiation dose reduction results in increased noise in the chest X-ray image. Thus, the purpose of this study was to evaluate the utility of fast non-local means (FNLM) filters to reduce radiation dose while maintaining sufficient image quality. 
Material and Methods: This study evaluates three filters (median, Wiener, and total variation) and a newly proposed filter (fast non-local means (FNLM)), which reduce image noise. A realistic anthropomorphic phantom is used to compare images acquired depending on positions such as anterior-posterior, lateral, and posterior-anterior, using a self-produced 3D printed lung nodule phantom. To evaluate image quality, we used the normalized noise power spectrum (NNPS), contrast to noise ratio (CNR), and coefficient of variation (COV) evaluation parameters. 
Results: The NNPS and COV were lowest and the CNR was highest with FNLM images. FNLM filter outperforms other compared filters in terms of noise reduction. 
Conclusion: Therefore, the use of an FNLM filter is recommended, because it reduces the radiation dose to a patient and thus minimizes the risk of cancer, while maintaining diagnostic quality.


Jina Shim
Department of Diagnostic Radiology, Severance Hospital, Seoul, Republic of Korea

Myonggeun Yoon
Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Youngjin Lee
Department of Radiological Science, Gachon University, 191, Hambakmoero, Yeonsu-gu, Incheon, Republic of Korea
Comparison of filter…
Comparison of filtered back projection with fast non-local m…
This study aimed to quantitatively analyze the effciency of the fast non-local means (FNLM) filter in increasing the nodule detection sensitivity in pediatric chest computed tomography (CT) using 3D-printed lung nodules. For that purpose, we compared filtered back projection (FBP) with FNLM filter with iterative reconstruction (IR) method. The performance of the proposed FNLM filter was compared through various quantitative evaluations by modeling the previously used noise reduction methods. When the FNLM filter was applied to the acquired FBP reconstruction method-based CT image, the coeffcient of variation and contrast-to-noise ratio values in the lung nodule region showed similar values to those of the IR method. In addition, it was demonstrated that the point spread function value that can evaluate sharpness can be improved by using the FNLM filter. In conclusion, the results of this study are expected to maximize the image quality and reduce the dose by fusing the CT image reconstructed by the FBP method and the FNLM filter with excellent characteristics.

Jina Shim
Department of Diagnostic Radiology, Severance Hospital, 50-1, Yonsei-ro, Seodaemun-gu, Seoul, Republic of Korea

Myonggeun Yoon*
Department of Bioengineering, Korea University, 145, Anam-ro, Seongbuk-gu, Seoul, Republic of Korea

Youngjin Lee*
Department of Radiological Science, Gachon University, 191, Hambakmoe-ro, Yeonsu-gu, Incheon, Republic of Korea 
Study of multistep D…
Study of multistep Dense U-Net-based automatic segmentation …
Background: Despite extensive efforts to obtain accurate segmentation of magnetic resonance imaging (MRI) scans of a head, it remains challenging primarily due to variations in intensity distribution, which depend on the equipment and parameters used. Purpose: The goal of this study is to evaluate the effectiveness of an automatic segmentation method for head MRI scans using a multistep Dense U-Net (MDU-Net) architecture. 
Methods: The MDU-Net-based method comprises two steps. The first step is to segment the scalp, skull, and whole brain from head MRI scans using a convolutional neural network (CNN). In the first step, a hybrid network is used to combine 2.5D Dense U-Net and 3D Dense U-Net structure. This hybrid network acquires logits in three orthogonal planes (axial, coronal, and sagittal) using 2.5D Dense U-Nets and fuses them by averaging. The resultant fused probability map with head MRI scans then serves as the input to a 3D Dense U-Net. In this process, different ratios of active contour loss and focal loss are applied. The second step is to segment the cerebrospinal fluid (CSF), white matter, and gray matter from extracted brain MRI scans using CNNs. In the second step, the histogram of the extracted brain MRI scans is standardized and then a 2.5D Dense U-Net is used to further segment the brain’s specific tissues using the focal loss. A dataset of 100 head MRI scans from an OASIS-3 dataset was used for training, internal validation, and testing, with ratios of 80%, 10%, and 10%, respectively. Using the proposed approach, we segmented the head MRI scans into five areas (scalp, skull, CSF, white matter, and gray matter) and evaluated the segmentation results using the Dice similarity coefficient (DSC) score, Hausdorff distance (HD), and the average symmetric surface distance (ASSD) as evaluation metrics.We compared these results with those obtained using the Res-U-Net, Dense U-Net, U-Net++, Swin-Unet, and H-Dense U-Net models. 
Results: The MDU-Net model showed DSC values of 0.933, 0.830, 0.833, 0.953, and 0.917 in the scalp, skull, CSF, white matter, and gray matter, respectively. The corresponding HD values were 2.37, 2.89, 2.13, 1.52, and 1.53 mm, respectively.The ASSD values were 0.50, 1.63, 1.28, 0.26, and 0.27 mm, respectively. Comparing these results with other models revealed that the MDU-Net model demonstrated the best performance in terms of the DSC values for the scalp, CSF, white matter, and gray matter. When compared with the H-Dense UNet model,which showed the highest performance among the other models,the MDU-Net model showed substantial improvements in the HD view, particularly in the gray matter region, with a difference of approximately 9%. In addition, in terms of the ASSD, the MDU-Net model outperformed the H-Dense U-Net model, showing an approximately 7% improvements in the white matter and approximately 9% improvements in the gray matter. 
Conclusion: Compared with existing models in terms of DSC, HD, and ASSD, the proposed MDU-Net model demonstrated the best performance on average and showed its potential to enhance the accuracy of automatic segmentation for head MRI scans.

Yongha Gi, Geon Oh, Hyeongjin Lim, Yousun Ko, Jinyoung Hong, Eunjun Lee, Sangmin Park, Taemin Kwak, Sangcheol Kim, Myonggeun Yoon*
Department of Bio-medical Engineering, Korea University, Seoul, Republic of Korea

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University, Seoul, Republic of Korea

Sangmin Park, Taemin Kwak, Sangcheol Kim, Myonggeun Yoon*
Field Cure Ltd., Seoul, Republic of Korea


Synergistic effect o…
Synergistic effect of TTF and 5-FU combination treatment on…
The present study investigated the therapeutic potential of combining tumor-treating fields (TTF), a novel cancer treatment modality that employs low-intensity, alternating electric fields, with 5-fluorouracil (5-FU), a standard chemotherapy drug used for treating pancreatic cancer. The HPAF-II and Mia-Paca II pancreatic cancer cell lines were treated with TTF, 5-FU, or their combination. Combination treatment produced a significantly greater inhibitory effect on cancer cell proliferation than each single modality. Furthermore, combination therapy induced a substantially higher rate of pancreatic cancer cell apoptosis and exhibited a synergistic effect in clonogenic assays. Additionally, combination treatment showed a greater inhibition of cancer cell migration and invasion than either TTF or 5-FU alone. In conclusion, these findings suggest that the synergistic properties of TTF and 5-FU result in greater therapeutic efficacy against pancreatic cancer cells than either modality alone and may improve survival rates in patients with pancreatic cancer.

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University, Seoul 02841, Republic of Korea

Eunjun Lee, Geon Oh, Yongha Gi, Myonggeun Yoon
Department of Bio-Medical Engineering, Korea University, Seoul 02841, Republic of Korea

Myonggeun Yoon
FieldCure Ltd., Seoul 02852, Republic of Korea
Development of a Dai…
Development of a Daily-Treatment Beam-Monitoring System Base…

This study describes the development of a simple method to assess inter-fractional deviations of delivered proton beams in treatment rooms. To monitor the treatment beam, we measured the field-by-field beam fluences by attaching the EBT3 film to the snout, followed by a simple constancy check based on comparisons between the reference beam fluences (acquired during the pre-treatment quality assurance process) and the test beam fluences (acquired during treatment). The feasibility of the proposed treatment beam-monitoring system was confirmed by evaluating 12 treatment fields for each of six patients (brain, liver, prostate, lung, cranial and spinal area, and head and neck). The constancy of the treatment beams was verified by using a gamma index analysis to compare three measurements per field with the reference beam fluence. On the basis of the 3%/3 mm criterion, the average gamma-index passing rates for all measurements were over 99.6%. These results suggest that the constancy of fractional proton beams delivered to patients in treatment rooms can be verified with EBT3 film-based proton-beam monitoring system that can be easily attached to the treatment nozzle and is cost effective.

Seonghoon Jeong

Proton Therapy Center, National Cancer Center, Goyang Korea

Myonggeun Yoon

Department of Bio-convergence Engineering, Korea University, Seoul, Korea

Kwangzoo Chung†

Department of Radiation Oncology, Samsung Medical Center, School of Medicine, Sungkyunkwan University, Seoul, Korea


Journal of the Korean Physical Society, Vol. 76, No. 8, April 2020, pp. 769773
Biophysical Model In…
Biophysical Model Including a Potentially Lethal Damage Repa…
The amount of potentially lethal damage repair (PLDR) is a significant factor in the process of modeling the survival curves of cells irradiated with fractionated carbon beams. Because the amount of PLDR generally depends on the features of the cells and the linear energy transfer (LET), the amount of PLDR of cells irradiated with fractionated carbon beams shows distinct differences from that of cells irradiated with X-rays. This study considered a new parameter dependent on the correlation between the PLDR trait (T) of the cells over a time interval (Δ) at the fractionated carbon beam irradiation. The survival curves of the cells irradiated with fractionated carbon beams n times were predicted using the ζ and the Ψ values from the delay assay. This study aims to overcome the barriers of traditional methods by developing a new survival curve model with new parameters based on an analysis of the PLDR traits of cells over time interval in fractionated carbon beam irradiation and to suggest a model that produces results significantly closer to the experimental data.

Eunae Choi and Myonggeun Yoon

Department of Bio-convergence Engineering, Korea University, Seoul 02841, Korea

Masao Suzuki and Naruhiro Matsufuji

National Institutes for Quantum and Radiological Science and Technology, Chiba 263-8555, Japan

Wonguyn Jung

Korea Institute of Radiological and Medical Science, Seoul 01812, Korea

Journal of the Korean Physical Society, Vol. 77, No. 2, July 2020, pp. 161∼167
Evaluating direct an…
Evaluating direct and indirect effects of low-energy electro…
Monte Carlo simulations can classify DNA damage into different types and predict the amount of energy deposited. Geant4-DNA was used to predict simple and complex DNA damage induced by irradiation of low-energy electrons at 0.1–50 keV. The number of molecules generated at different energy levels of radiation was analyzed after observing the gradual changes in the level of water radiolysis. ADNA model was used to categorize direct damage according to the location of strand breaks at the atomic level. The parameters of energy threshold (minimum amount of energy needed to break DNA strands) and 10 base pairs (maximum distance that separates two strand breaks) were set. All instances of water radiolysis including the main OH radical occurred most frequently at 1 keV followed by at 1.5 and 0.5 keV. Direct strand breaks most commonly occurred at 0.5 keV followed by at 0.3 keV. Finally, most of strand breaks occurred more frequently at 0.5 keV than at 0.3 keV. The computational measure-ment results for indirect and direct effects of irradiation depend on the type of simulation code and the DNA model used. Values used in Geant4 (physics list, chemical interaction time and energy threshold)may also influence the results.

Eunae Choi and Myong Geun Yoon

Department of Bio Convergence Engineering, Korea University, Seongbuk-gu, Republic of Korea

Kwon Su Chon

Catholic University of Daegu, Gyeongsan, Republic of Korea


RADIATION EFFECTS & DEFECTS IN SOLIDS2020, VOL. 175, NOS. 11–12, 1042–1051
Thymidine decreases …
Thymidine decreases the DNA damage and apoptosis caused by t…

Background

Tumor-treating fields (TTFields) is an emerging non-invasive cancer-treatment modality using alternating electric fields with low intensities and an intermediate range of frequency. TTFields affects an extensive range of charged and polarizable cellular factors known to be involved in cell division. However, it causes side-effects, such as DNA damage and apoptosis, in healthy cells.

Objective

To investigate whether thymidine can have an effect on the DNA damage and apoptosis, we arrested the cell cycle of human glioblastoma cells (U373) at G1/S phase by using thymidine and then exposed these cells to TTFields.

Methods

Cancer cell lines and normal cell (HaCaT) were arrested by thymidine double block method. Cells were seeded into the gap of between the insulated wires. The exposed in alternative electric fields at 120 kHz, 1.2 V/cm. They were counted the cell numbers and analyzed for cancer malignant such as colony formation, Annexin V/PI staining, γH2AX and RT-PCR.

Results

The colony-forming ability and DNA damage of the control cells without thymidine treatment were significantly decreased, and the expression levels of BRCA1, PCNA, CDC25C, and MAD2 were distinctly increased. Interestingly, however, cells treated with thymidine did not change the colony formation, apoptosis, DNA damage, or gene expression pattern.

Conclusions

These results demonstrated that thymidine can inhibit the TTFields-caused DNA damage and apoptosis, suggesting that combining TTFields and conventional treatments, such as chemotherapy, may enhance prognosis and decrease side effects compared with those of TTFields or conventional treatments alone.


Hyesun Jeong, Sunghoi Hong

School of Biosystems and Biomedical Sciences, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea

Hyesun Jeong, Sunghoi Hong

Department of Public Health Science, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea

Yunhui Jo, Myonggeun Yoon

Department of Bio- Convergence Engineering, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea


Genes Genom 43, 995–1001 (2021)

Sensitivity of radio…
Sensitivity of radio-photoluminescence glass dosimeters to a…
Background 
This study investigated the effect of accumulated doses on radio-photoluminescence glass dosimeters (RPLGDs) from measurements involving mega-voltage photons. 
Methods 
Forty-five commercially available RPLGDs were irradiated to estimate their dose responses. Photon beams of 6, 10, and 15 MV were irradiated onto the RPLGDs inside a phantom, which were divided into five groups with different doses and energies. Groups 1 and 2 were irradiated at 1, 5, 10, 50, and 100 Gy in a sequential manner; Group 3 was irradiated 10 times with a dose of 10 Gy; and Groups 4 and 5 followed the same method as that of Group 3, but with doses of 50 Gy and 100 Gy, respectively. Each device was subjected to a measurement reading procedure each time irradiation. 
Results 
For the annealed Group 1, RPLGD exhibited a linearity response with variance within 5%. For the non-annealed Group 2, readings demonstrated hyperlinearity at 6 MV and 10 MV, and linearity at 15 MV. Following the 100 Gy irradiation, the readings for Group 2 were 118.7 ± 1.9%, 112.2 ± 2.7%, and 101.5 ± 2.3% at 6, 10, and 15 MV, respectively. For Groups 3, 4, and 5, the responsiveness of the RPLGDs gradually decreased as the number of repeated irradiations increased. The percentage readings for the 10th beam irradiation with respect to the readings for the primary beam irradiation were 84.6 ± 1.9%, 87.5 ± 2.4%, and 93.0 ± 3.0% at 6 MV, 10 MV, and 15 MV, respectively. 
Conclusions 
The non-annealed RPLGD response to dose was hyperlinear for the 6 MV and 10 MV photon beams but not for the 15 MV photon beam. Additionally, the annealed RPLGD exhibited a fading phenomenon when the measurement was repeated several times and demonstrated a relatively large fading effect at low energies than at high energies.

Dong Wook Kim, Han-Back Shin, Min-Joo Kim, Yu-Yun Noh, Hojae Kim, Min Cheol Han, Jihun Kim, Su Chul Han, Kyung Hwan Chang, Hojin Kim, Kwangwoo Park, Jinsung Kim
Department of Radiation Oncology, Yonsei Cancer Center, Yonsei University College of Medicine, Seoul, Korea
Jiwon Sung, Jaeman Son
Department of Radiation Oncology, Seoul National University Hospital, Seoul, Korea
Myonggeun Yoon 
Department of Bio-Convergence Engineering, Korea University, Seoul, Korea
Dongho Shin
Proton Therapy Center, National Cancer Center, Goyang, Korea

PLOS ONE | https://doi.org/10.1371/journal.pone.0234829 December 3, 2020
Feasibility Study of…
Feasibility Study of Beam Angle Optimization for Proton Trea…
This study describes a method that uses a genetic algorithm to select optimal beam angles in proton therapy and evaluates the effectiveness of the proposed algorithm in actual patients. In the use of the genetic algorithm to select the optimal angle, a gene represents the angle of each field and a chromosome represents the combination of beam angles. The fitness of the genetic algorithm, which represents the suitability of the chromosome to the solution, was quantified by using the dose distribution. The weighting factors of the organs used for fitness were obtained from clinical data through logistic regression, reflecting the dose characteristics of actual patients. Genetic operations, such as selection, crossover, mutation, and replacement, were used to modify the population and were repeated until an evaluation based on fitness reached the termination criterion. The proposed genetic algorithm was tested by assessing its ability to select optimal beam angles in three patients with liver cancer. The optimal results for fitness, planning target volume (PTV), normal liver, and skin in the population were compared with the clinical treatment plans, a process that took an average of 36.8 minutes. The dose-volume histograms (DVHs) and the fitness of the genetic algorithm plans did not differ significantly from the actual treatment plans. These findings indicate that the proposed genetic algorithm can automatically generate proton treatment plans with the same quality as actual clinical treatment plans.

Jaehyeon Seo, Yunhui Jo, Sunyoung Moon and Myonggeun Yoon

Department of Bio-convergence Engineering, Korea University, Seoul 02841, Korea

Sung Hwan Ahn, Boram Lee and Kwangzoo Chung†

Department of Radiation Oncology, Samsung Medical Center, Sungkyunkwan University School of Medicine, Seoul 06351, Korea

Seonghoon Jeong

Department of Proton Therapy Center, National Cancer Center, Goyang 10408, Korea

Journal of the Korean Physical Society, Vol. 77, No. 4, August 2020, pp. 312∼316
A new real-time pers…
A new real-time personal dosimeter with position monitoring …
Personal dosimeters are used to measure the amount of radiation exposure in individual radiation workers. We aimed to replace existing personal dosimeters and evaluate a real-time scintillator-based dosimeter by monitoring its radiation dose and checking the location exposed to radiation in the workspace. The developed dosimeter measured the radiation dose based on a scintillating fiber (SF) bundle, and comprised a silicon photomultiplier (SiPM), ultra-wide-band (UWB)-based location detecting system, and Bluetooth system. The SF bundle was exposed to radiation-emitted light, and the photons were amplified and converted to electrical signals through the SiPM. These signals were transferred to the user through the Bluetooth system and monitored. To evaluate the feasibility of this mechanism as a dosimeter, we performed characteristic tests, such as dose linearity, dependence on dose rate, energy, exposed angle, and location coordinate mapping. Also, the dose distribution formed in circles around the iso-center was measured to confirm the feasibility of monitoring the exposure dose and location and to enable the radiation worker to move freely in a workspace. We confirmed dose linearity, independence from energy and angle, and accuracy of location monitoring in our device. The user’s locations were measured with a difference of 6 cm and 4.8 cm on the x- and the y-axes, respectively. The measured doses on our developed dosimeter were 62.7, 32.3, 21.0, and 15.4 mSv at distances of 50, 100, 150, and 200 cm from the iso-center. In other words, all measured doses at several points showed an error within 5% as compared to doses provided by the conventional pocket dosimeter. These results show that the developed SF-based dosimeter is advantageous in monitoring the exposure dose and location in real time, and has significant potential as a new personal dosimeter for radiation workers.

Sun Young Moon, Myonggeun Yoon

Department of Bio-Convergence Engineering, Korea University, Seoul, South Korea

Sun-Young Kim, Dongho Shin

Proton Therapy Center, National Cancer Center, Goyang, South Korea


J. Korean Phys. Soc. 78, 1133–1140 (2021).

Clinical application…
Clinical application of a gantry-attachable plastic scintill…

Purpose: The entrance beam fluence of therapeutic proton scanning beams can be monitored using a gantry-attachable plastic scintillating plate (GAPSP). This study evaluated the clinical application of the GAPSP using amethod that measures intensity modulated proton therapy (IMPT) beams for patient treatment.

Methods: IMPT beams for the treatment of nine patients, at sites that included the spine, head and neck, pelvis, and lung, were measured using the GAPSP, composed of an EJ-212 plastic scintillator and a CMOS camera. All energy layers distinguished by the GAPSP were accumulated to determine the dose distribution of the treatment field. The evaluated fields were compared with reference dose maps verified by quality assurance.

Results: Comparison of dose distributions of evaluation treatment fields with reference dose distributions showed that the 3%/1 mm average gamma passing rate was 96.4%, independent of the treatment site, energy range and field size. When dose distributions were evaluated using the same criteria for each energy layer, the average gamma passing rate was 91.7%.

  Conclusions: The GAPSP is a suitable, low-cost method for monitoring pencil beam scanning proton therapy, especially for non-spot scanning or additional collimation. The GAPSP can also estimate the treatment beam by the energy layer, a feature not common to other proton dosimetry tool.


Seonghoon Jeong

Proton Therapy Center, National Cancer Center, Goyang, Republic of Korea

Myonggeun Yoon , Jaehyeon Seo

Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Kwangzoo Chung, , Sung Hwan Ahn , Boram Lee,

Department of Radiation Oncology, Samsung Medical Center, Seoul, Republic of Korea

Physica Medica 77(2020)181-186