Clinical application of a gantry-attachable plastic scintillating plate dosimetry system in pencil beam scanning proton therapy beam monitoring
 글쓴이 : 관리자
조회 : 428  

Purpose: The entrance beam fluence of therapeutic proton scanning beams can be monitored using a gantry-attachable plastic scintillating plate (GAPSP). This study evaluated the clinical application of the GAPSP using amethod that measures intensity modulated proton therapy (IMPT) beams for patient treatment.

Methods: IMPT beams for the treatment of nine patients, at sites that included the spine, head and neck, pelvis, and lung, were measured using the GAPSP, composed of an EJ-212 plastic scintillator and a CMOS camera. All energy layers distinguished by the GAPSP were accumulated to determine the dose distribution of the treatment field. The evaluated fields were compared with reference dose maps verified by quality assurance.

Results: Comparison of dose distributions of evaluation treatment fields with reference dose distributions showed that the 3%/1 mm average gamma passing rate was 96.4%, independent of the treatment site, energy range and field size. When dose distributions were evaluated using the same criteria for each energy layer, the average gamma passing rate was 91.7%.

  Conclusions: The GAPSP is a suitable, low-cost method for monitoring pencil beam scanning proton therapy, especially for non-spot scanning or additional collimation. The GAPSP can also estimate the treatment beam by the energy layer, a feature not common to other proton dosimetry tool.


Seonghoon Jeong

Proton Therapy Center, National Cancer Center, Goyang, Republic of Korea

Myonggeun Yoon , Jaehyeon Seo

Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Kwangzoo Chung, , Sung Hwan Ahn , Boram Lee,

Department of Radiation Oncology, Samsung Medical Center, Seoul, Republic of Korea

Physica Medica 77(2020)181-186