Sensitivity of radio…
Sensitivity of radio-photoluminescence glass dosimeters to a…
Background 
This study investigated the effect of accumulated doses on radio-photoluminescence glass dosimeters (RPLGDs) from measurements involving mega-voltage photons. 
Methods 
Forty-five commercially available RPLGDs were irradiated to estimate their dose responses. Photon beams of 6, 10, and 15 MV were irradiated onto the RPLGDs inside a phantom, which were divided into five groups with different doses and energies. Groups 1 and 2 were irradiated at 1, 5, 10, 50, and 100 Gy in a sequential manner; Group 3 was irradiated 10 times with a dose of 10 Gy; and Groups 4 and 5 followed the same method as that of Group 3, but with doses of 50 Gy and 100 Gy, respectively. Each device was subjected to a measurement reading procedure each time irradiation. 
Results 
For the annealed Group 1, RPLGD exhibited a linearity response with variance within 5%. For the non-annealed Group 2, readings demonstrated hyperlinearity at 6 MV and 10 MV, and linearity at 15 MV. Following the 100 Gy irradiation, the readings for Group 2 were 118.7 ± 1.9%, 112.2 ± 2.7%, and 101.5 ± 2.3% at 6, 10, and 15 MV, respectively. For Groups 3, 4, and 5, the responsiveness of the RPLGDs gradually decreased as the number of repeated irradiations increased. The percentage readings for the 10th beam irradiation with respect to the readings for the primary beam irradiation were 84.6 ± 1.9%, 87.5 ± 2.4%, and 93.0 ± 3.0% at 6 MV, 10 MV, and 15 MV, respectively. 
Conclusions 
The non-annealed RPLGD response to dose was hyperlinear for the 6 MV and 10 MV photon beams but not for the 15 MV photon beam. Additionally, the annealed RPLGD exhibited a fading phenomenon when the measurement was repeated several times and demonstrated a relatively large fading effect at low energies than at high energies.

Dong Wook Kim, Han-Back Shin, Min-Joo Kim, Yu-Yun Noh, Hojae Kim, Min Cheol Han, Jihun Kim, Su Chul Han, Kyung Hwan Chang, Hojin Kim, Kwangwoo Park, Jinsung Kim
Department of Radiation Oncology, Yonsei Cancer Center, Yonsei University College of Medicine, Seoul, Korea
Jiwon Sung, Jaeman Son
Department of Radiation Oncology, Seoul National University Hospital, Seoul, Korea
Myonggeun Yoon 
Department of Bio-Convergence Engineering, Korea University, Seoul, Korea
Dongho Shin
Proton Therapy Center, National Cancer Center, Goyang, Korea

PLOS ONE | https://doi.org/10.1371/journal.pone.0234829 December 3, 2020
Utility of fast non-…
Utility of fast non-local means (FNLM) filter for detection …

Purpose: This study was aimed to evaluate the utility based on imaging quality of the fast non-local means (FNLM) filter in diagnosing lung nodules in pediatric chest computed tomography (CT).

Methods: We retrospectively reviewed the chest CT reconstructed with both filtered back projection (FBP) and iterative reconstruction (IR) in pediatric patients with metastatic lung nodules. After applying FNLM filter with six h values (0.0001, 0.001, 0.01, 0.1, 1, and 10) to the FBP images, eight sets of images including FBP, IR, and FNLM were analyzed. The image quality of the lung nodules was evaluated objectively for coefficient of variation (COV), contrast to noise ratio (CNR), and point spread function (PSF), and subjectively for noise, sharpness, artifacts, and diagnostic acceptability.

Results: The COV was lowest in IR images and decreased according to increasing h values and highest with FBP images (P < 0.001). The CNR was highest with IR images, increased according to increasing h values and lowest with FBP images (P < 0.001). The PSF was lower only in FNLM filter with h value of 0.0001 or 0.001 than in IR images (P < 0.001). In subjective analysis, only images of FNLM filter with h value of 0.0001 or 0.001 rarely showed unacceptable quality and had comparable results with IR images. There were less artifacts in FNLM images with h value of 0.0001 compared with IR images (p < 0.001).

Conclusion: FNLM filter with h values of 0.0001 allows comparable image quality with less artifacts compared


Jina Shim, Myonggeun Yoon*

Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Department of Diagnostic Radiology, Severance Hospital, Seoul, Republic of Korea

Mi-Jung Lee* 

Department of Radiology and Research Institute of Radiological Science, Severance Children’s Hospital, Yonsei University, College of Medicine, Seoul, Republic of Korea

Youngjin Lee

Department of Radiological Science, Gachon University, Incheon, Republic of Korea


Physica Medica 81 (2021) 52-89

A new real-time pers…
A new real-time personal dosimeter with position monitoring …
Personal dosimeters are used to measure the amount of radiation exposure in individual radiation workers. We aimed to replace existing personal dosimeters and evaluate a real-time scintillator-based dosimeter by monitoring its radiation dose and checking the location exposed to radiation in the workspace. The developed dosimeter measured the radiation dose based on a scintillating fiber (SF) bundle, and comprised a silicon photomultiplier (SiPM), ultra-wide-band (UWB)-based location detecting system, and Bluetooth system. The SF bundle was exposed to radiation-emitted light, and the photons were amplified and converted to electrical signals through the SiPM. These signals were transferred to the user through the Bluetooth system and monitored. To evaluate the feasibility of this mechanism as a dosimeter, we performed characteristic tests, such as dose linearity, dependence on dose rate, energy, exposed angle, and location coordinate mapping. Also, the dose distribution formed in circles around the iso-center was measured to confirm the feasibility of monitoring the exposure dose and location and to enable the radiation worker to move freely in a workspace. We confirmed dose linearity, independence from energy and angle, and accuracy of location monitoring in our device. The user’s locations were measured with a difference of 6 cm and 4.8 cm on the x- and the y-axes, respectively. The measured doses on our developed dosimeter were 62.7, 32.3, 21.0, and 15.4 mSv at distances of 50, 100, 150, and 200 cm from the iso-center. In other words, all measured doses at several points showed an error within 5% as compared to doses provided by the conventional pocket dosimeter. These results show that the developed SF-based dosimeter is advantageous in monitoring the exposure dose and location in real time, and has significant potential as a new personal dosimeter for radiation workers.

Sun Young Moon, Myonggeun Yoon

Department of Bio-Convergence Engineering, Korea University, Seoul, South Korea

Sun-Young Kim, Dongho Shin

Proton Therapy Center, National Cancer Center, Goyang, South Korea


J. Korean Phys. Soc. 78, 1133–1140 (2021).

Thymidine decreases …
Thymidine decreases the DNA damage and apoptosis caused by t…

Background

Tumor-treating fields (TTFields) is an emerging non-invasive cancer-treatment modality using alternating electric fields with low intensities and an intermediate range of frequency. TTFields affects an extensive range of charged and polarizable cellular factors known to be involved in cell division. However, it causes side-effects, such as DNA damage and apoptosis, in healthy cells.

Objective

To investigate whether thymidine can have an effect on the DNA damage and apoptosis, we arrested the cell cycle of human glioblastoma cells (U373) at G1/S phase by using thymidine and then exposed these cells to TTFields.

Methods

Cancer cell lines and normal cell (HaCaT) were arrested by thymidine double block method. Cells were seeded into the gap of between the insulated wires. The exposed in alternative electric fields at 120 kHz, 1.2 V/cm. They were counted the cell numbers and analyzed for cancer malignant such as colony formation, Annexin V/PI staining, γH2AX and RT-PCR.

Results

The colony-forming ability and DNA damage of the control cells without thymidine treatment were significantly decreased, and the expression levels of BRCA1, PCNA, CDC25C, and MAD2 were distinctly increased. Interestingly, however, cells treated with thymidine did not change the colony formation, apoptosis, DNA damage, or gene expression pattern.

Conclusions

These results demonstrated that thymidine can inhibit the TTFields-caused DNA damage and apoptosis, suggesting that combining TTFields and conventional treatments, such as chemotherapy, may enhance prognosis and decrease side effects compared with those of TTFields or conventional treatments alone.


Hyesun Jeong, Sunghoi Hong

School of Biosystems and Biomedical Sciences, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea

Hyesun Jeong, Sunghoi Hong

Department of Public Health Science, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea

Yunhui Jo, Myonggeun Yoon

Department of Bio- Convergence Engineering, Korea University, 145 Anam-ro, Seongbuk-gu, Seoul, 02841, Republic of Korea


Genes Genom 43, 995–1001 (2021)

Inter fractional ent…
Inter fractional entrance dose monitoring as quality assuran…
Introduction: This study describes a simple method of inter‑fractional photon beam monitoring to measure the entrance dose of radiation treatment using Gafchromic EBT3 film. 
Materials and Methods: The film was placed at the center of a 1‑cm thick phantom shaped like a block tray and fixed on the accessory tray of the gantry. The entrance dose was measured following the placement of the film in the accessory tray. The dose distribution calculated with the treatment planning system was compared with the dose distribution on the irradiated EBT3 films. The effectiveness of this methodology, as determined by gamma passing rates, was evaluated for the 22 fields of eight three‑dimensional conformal radiotherapy (3D‑CRT) plans and the 41 fields of nine intensity‑modulated radiotherapy (RT) plans. The plans for three‑dimensional conformal RT included treatments of the rectum, liver, breast, and head and neck, whereas the plans for intensity‑modulated RT included treatments of the liver, brain, and lung. 
Results: The gamma passing rates for 3D‑CRT ranged from 96.4% to 99.5%, with the mean gamma passing rate for 22 fields being 98.0%. The gamma passing rate for intensity‑modulated RT ranged from 96.1% to 98.9%, with the mean gamma passing rate for 41 fields being 97.7%. All gamma indices were over the 95% tolerance level. 
Conclusions: The methodology described in this study, based on Gafchromic EBT3 film, can be utilized for inter‑fractional entrance dose monitoring as quality assurance during RT. Clinical application of this method to patients can verify the accuracy of beam delivery in the treatment room.

Sun Young Moon
Department of Bio‑convergence Engineering, Korea University, Seoul
Proton Therapy Center, National Cancer Center, Goyang, Korea
Yunhui Jo, Jaehyeon Seo, Myonggeun Yoon
Department of Bio‑convergence Engineering, Korea University, Seoul 
Dongho Shin,  
Proton Therapy Center, National Cancer Center, Goyang, Korea

Journal of Cancer Research and Therapeutics - Volume 18 - Issue 4 - July-September 2022: 1152-1158
The combination of t…
The combination of tumor treating fields and hyperthermia ha…

Glioblastoma multiforme (GBM), the most common type of brain tumor, is a very aggressive and treatment-refractory cancer, with a 5-year survival rate of approximately 5%. Hyperthermia (HT) and tumor treating fields (TTF) therapy have been used to treat cancer, either alone or in combination with other treatment methods. Both treatments have been reported to increase the efficacy of other treatment techniques and to improve patient prognosis. The present study evaluated the therapeutic effects of combining HT and TTF on GBM cell lines. Cells were subjected to HT, TTF, HT+TTF, or neither treatment, followed by comparisons of cell proliferation, apoptosis, migration and invasiveness. Clonogenic assays showed that the two treatments had a synergistic effect. The levels of cleaved PARP and cleaved caspase-3 were higher and apoptosis was increased in cells treated with HT+TTF than in cells treated with HT or TTF alone. In addition, HT+TTF showed greater inhibition of GBM cell migration and invasiveness and greater downregulation of STAT3 than either HT or TTF alone. The stronger anticancer effect of HT+TTF suggested that this combination treatment can increase the survival rate of patients with difficult-to-treat cancers such as GBM.


Yunhui Jo, Young In Han

Institute of Global Health Technology (IGHT), Korea University, Seoul, Republic of Korea

Eunjun Lee, Geon Oh, Heehun Sung, Yongha Gi, Hyunwoo Kim, Sangmin Park

Department of Biomedical Engineering, Korea University, Seoul, Republic of Korea

Jaehyeon Seo

Department of Bioconvergence Engineering, Korea University, Seoul, Republic of Korea 

Myonggeun Yoon

Department of Biomedical Engineering, Korea University, Seoul, Republic of Korea

Department of Bioconvergence Engineering, Korea University, Seoul, Republic of Korea


Am J Cancer Res. 2022; 12(3): 1423–1432.

Development of a dos…
Development of a dosimetry system for therapeutic X-rays usi…
Purpose: To evaluate the dosimetric characteristics and applications of a dosimetry system composed of a flexible amorphous silicon thin-film solar cell and scintillator screen (STFSC-SS) for therapeutic X-rays. 
Methods: The real-time dosimetry system was composed of a flexible a Si thin-film solar cell (0.2-mm thick), a scintillator screen to increase its efficiency, and an electrometer to measure the generated charge. The dosimetric characteristics of the developed system were evaluated including its energy dependence, dose linearity, and angular dependence. Calibration factors for the signal measured by the system and absorbed dose-to-water were obtained by setting reference conditions. The application and correction accuracy of the developed system were evaluated by comparing the absorbed dose-to water measured using a patient treatment beam with that measured using the ion chamber. 
Results: The responses of STFSC-SS to energy, field size, depth, and sourceto-surface distance (SSD) were more dependent on measurement conditions than were the responses of the ion chamber, although the former dependence was due to the scintillator screen, not the solar cell. The signals of STFSC-SS were also dependent on dose rate, while the responses of solar cell alone and scintillator screen were not dependent on dose rate. The scintillator screen reduced the output of solar cell at 6 and 15 MV by 0.60 and 0.55%, respectively. The different absorbed dose-to-water measured using STFSC-SS for patient treatment beam differed by 0.4% compared to those measured using the ionization chamber. The uncertainties of the developed system for 6 and 15 MV photon beams were 1.8 and 1.7%, respectively, confirming the accuracy and applicability of this system. 
Conclusions: The thin-film solar cell-based detector developed in this study can accurately measure absorbed dose-to-water. The increased signal resulting from the use of the scintillator screen is advantageous for measuring low doses and stable signal output. In addition, this system is flexible, making it applicable to curved surfaces, such as a patient’s body, and is cost-effective.

Seonghoon Jeong, Wonjoong Cheon, Dongho Shin, Young Kyung Lim, Jonghwi Jeong, Haksoo Kim, Se Byeong Lee 
Proton Therapy Center, National Cancer Center, Goyang, Korea
Myonggeun Yoon
Department of Bioconvergence Engineering, Korea University, Seoul, Korea

Med Phys. 2022;49:4768–4779
Technical note: Eval…
Technical note: Evaluation of methods for reducing edge curr…

Background: Tumor-treating fields (TTFields) therapy is increasingly utilized clinically because of its demonstrated efficacy in cancer treatment. However, the risk of skin burns must still be reduced to improve patient safety and posttreatment quality of life.

Purpose: The purpose of this study was to evaluate the methods of constructing electrode arrays that reduce current density exceeding threshold values, which can cause skin burns during TTFields therapy.

Methods: Electrode and body models were generated using COMSOL software. The body model had the dielectric properties of the scalp. The average current density beneath the central region of the electrode was maintained at 31 mA/cm2 RMS. The deviations in current density at the edges of the electrode were reduced by three methods:adjustment of the ceramic thickness ratio of the center to the edge from 1/5 to 4/5, adjustment of the radius of the metal plate from 5.0 to 8.0 mm, and insertion of an insulator of width 0.5 to 2 mm at the edge.

Results: While using a single circular electrode, adjustment of the ceramic thickness ratio, adjustment of the metal plate radius, and insertion of an insulator near the edge reduced the deviations of current density by 14.6%, 67.7%, and 75.3%, respectively. Similarly, while using circular electrode arrays, inserting an insulator at the edge of each electrode reduced the deviations of current density significantly, from 8.62 to 2.40 mA/cm2.

Conclusions: Insertion of an insulator at the edge of each electrode was found to be the most effective method of attaining uniform current density distribution beneath the electrode, thereby lowering the risk of adverse effects of TTFields therapy.


Heehun Sung Geon Oh, Yongha Gi, Jaehyeon Seo

Department of Bioengineering, Korea University, Seoul,Republic of Korea

Yunhui Jo 

Institute of Global Health Technology (IGHT), Korea University, Seoul,Republic of Korea

Hyunwoo KimSangmin Park, Myonggeun Yoon

Department of Bioengineering, Korea University, Seoul,Republic of Korea

FieldCure Ltd, Seoul,Republic of Korea


Med Phys. 2022;49:4837–4844.

Feasibility of a qua…
Feasibility of a quality assurance system for electromagneti…
This study was designed to evaluate the effectiveness of a newly developed quality assurance system for electromagnetic field therapy called tumor treating fields therapy, which uses an alternating electric field to treat cancer based on an intermediate frequency of 100–300 kHz. The quality assurance system for electromagnetic field therapy consisted of a water phantom, a probe, and a digital data acquisition (DAQ) board. Low intensity alternating electric fields (200 kHz, 0–.1 V/cm) were created within the water phantom using a function generator and a high voltage amplifier. The electric potential formed inside the water phantom was measured using the probe and DAQ board. The electric field intensity was derived by measuring the electric potential at the 190 points (19×10 cm2) of the midplane. Accuracy was evaluated by gamma index analysis, which compared the measured electric field and the simulation result. The mean difference between the simulation result and the measured electric potentials within the water phantom was 0.31 V. The gamma passing rate for the tolerance levels of 0.5 V/5 mm was 95.5% for electric potential comparison showing good agreement between simulation and experimental results. The mean difference between the electric field distribution within the water phantom and the simulated values was 0.09 V/cm and the gamma passing rate for the tolerance levels of (0.2 V/cm)/5 mm was ~ 97.3%. These results confirmed the feasibility of the quality assurance system for electromagnetic field therapy.

Heehun Sung, Jaehyeon Seo, Geon Oh, Yongha Gi, Hyunwoo Kim & Myonggeun Yoon

Department of Bioengineering, Korea University, Seoul, Republic of Korea

Jong Hyun Kim & Myonggeun Yoon

FieldCure Ltd, Seoul, Republic of Korea

Yunhui Jo

Institute of Global Health Technology, Korea University, Seoul, Republic of Korea


J. Korean Phys. Soc. 81, 1029–1038 (2022). 

Feasibility of a met…
Feasibility of a method for optimizing the electrode array s…
The present study investigated electrode array structures that maximize the therapeutic electric field intensity to tumors with different shapes and locations, while minimizing electric field intensity to the surrounding organs at risk (OARs). A human body phantom model was created from magnetic resonance images of a patient and divided into regions including a tumor and OARs. The shapes and sizes of the electrode arrays were altered for tumors differing in shape and location, and these electrode array structures were tested in the phantom. Use of a conformal electrode array based on the shape of the tumor maintained therapeutic electric field intensity to the tumor while reducing electric field intensity to the surrounding OARs by approximately 18%. Although the electric field intensity delivered to the tumor was proportional to the size of the electrode array, it was saturated at a critical area. Simulation results showed that optimal sizes of electrode arrays for specific tumors located at depths of 2 cm, 4 cm and 6 cm were 91, 273 and 830 cm2, respectively, indicating that the optimal size of the electrode array is proportional to the depth of tumor in the phantom. These results suggest that a tumor location-dependent optimal ratio between the size of the electrode array and the size of the individual electrodes could be calculated. In summary, this study indicated that customizing the electrode array structure to individual tumors can markedly increase the electric field intensity delivered to the tumor while minimizing the intensity delivered to OARs.

Geon Oh, Yongha Gi, Heehun Sung, Hyunwoo Kim, Jaehyeon Seo & Myonggeun Yoon

Department of Bioengineering, Korea University, Seoul, 02841, Republic of Korea

Myonggeun Yoon

FieldCure Ltd, Seoul, 02481, Republic of Korea

Yunhui Jo

Institute of Global Health Technology, Korea University, Seoul, 02841, Republic of Korea


J. Korean Phys. Soc. 81, 1020–1028 (2022).

Potential prognostic…
Potential prognostic factor in alternating electric fields t…
The present study aimed to determine a method for estimating a potential prognostic factor in alternating electric fields for the treatment of solid tumors based on cell survival curves that evaluate cell proliferation capability. In AGS, B16F10, U373, and HPAF-II cancer cell lines, the proportional relationships of the electric field magnitude and the duration of application with the proliferation of cancer cell lines was identified by in vitro alternating current electric field experiments performed under various conditions. A prognostic factor applicable to alternating electric field therapy was developed by identifying proportional relationships of the electric field magnitude and the duration of application with the proliferation of the four cancer cell lines. Through the experimental results, the absorbed energy in tissue has been suggested as a potential prognostic factor in alternating electric field therapy. The absorbed energy in tissue can be used as a reference to quantify the inhibition of cell proliferation related to control, enabling systematic assessment of alternating electric field therapy which, to date, has not been possible.

Geon Oh, Yongha Gi, Heehun Sung, Jaehyun Seo, Hyunwoo Kim

Department of Bioengineering, Korea University, Seoul 02841,Republic of Korea 

Yunhui Jo

Institute of Global Health Technology(IGHT), Korea University, Seoul 02841, Republic of Korea 

Jaemin Lee

Department of Internal Medicine, Anam Hospital, Korea University College of Medicine, Seoul 02841, Republic of Korea 

Myonggeun Yoon

Department of Bioengineering, Korea University, Seoul 02841,Republic of Korea 

FieldCure Ltd.,Seoul 02852, Republic of Korea



AIPAdvances12,095311(2022)
Application of Fast …
Application of Fast Non-Local Denoising Approach in Digital …
Introduction: Chest X-ray imaging has become the most commonly used, as it is the primary method for lung cancer screening during medical check-ups. The radiation dose should be minimized to ensure that the patients are not overexposed to radiation. However, radiation dose reduction results in increased noise in the chest X-ray image. Thus, the purpose of this study was to evaluate the utility of fast non-local means (FNLM) filters to reduce radiation dose while maintaining sufficient image quality. 
Material and Methods: This study evaluates three filters (median, Wiener, and total variation) and a newly proposed filter (fast non-local means (FNLM)), which reduce image noise. A realistic anthropomorphic phantom is used to compare images acquired depending on positions such as anterior-posterior, lateral, and posterior-anterior, using a self-produced 3D printed lung nodule phantom. To evaluate image quality, we used the normalized noise power spectrum (NNPS), contrast to noise ratio (CNR), and coefficient of variation (COV) evaluation parameters. 
Results: The NNPS and COV were lowest and the CNR was highest with FNLM images. FNLM filter outperforms other compared filters in terms of noise reduction. 
Conclusion: Therefore, the use of an FNLM filter is recommended, because it reduces the radiation dose to a patient and thus minimizes the risk of cancer, while maintaining diagnostic quality.


Jina Shim
Department of Diagnostic Radiology, Severance Hospital, Seoul, Republic of Korea

Myonggeun Yoon
Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea

Youngjin Lee
Department of Radiological Science, Gachon University, 191, Hambakmoero, Yeonsu-gu, Incheon, Republic of Korea
Synergistic effect o…
Synergistic effect of TTF and 5-FU combination treatment on…
The present study investigated the therapeutic potential of combining tumor-treating fields (TTF), a novel cancer treatment modality that employs low-intensity, alternating electric fields, with 5-fluorouracil (5-FU), a standard chemotherapy drug used for treating pancreatic cancer. The HPAF-II and Mia-Paca II pancreatic cancer cell lines were treated with TTF, 5-FU, or their combination. Combination treatment produced a significantly greater inhibitory effect on cancer cell proliferation than each single modality. Furthermore, combination therapy induced a substantially higher rate of pancreatic cancer cell apoptosis and exhibited a synergistic effect in clonogenic assays. Additionally, combination treatment showed a greater inhibition of cancer cell migration and invasion than either TTF or 5-FU alone. In conclusion, these findings suggest that the synergistic properties of TTF and 5-FU result in greater therapeutic efficacy against pancreatic cancer cells than either modality alone and may improve survival rates in patients with pancreatic cancer.

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University, Seoul 02841, Republic of Korea

Eunjun Lee, Geon Oh, Yongha Gi, Myonggeun Yoon
Department of Bio-Medical Engineering, Korea University, Seoul 02841, Republic of Korea

Myonggeun Yoon
FieldCure Ltd., Seoul 02852, Republic of Korea
Comparison of filter…
Comparison of filtered back projection with fast non-local m…
This study aimed to quantitatively analyze the effciency of the fast non-local means (FNLM) filter in increasing the nodule detection sensitivity in pediatric chest computed tomography (CT) using 3D-printed lung nodules. For that purpose, we compared filtered back projection (FBP) with FNLM filter with iterative reconstruction (IR) method. The performance of the proposed FNLM filter was compared through various quantitative evaluations by modeling the previously used noise reduction methods. When the FNLM filter was applied to the acquired FBP reconstruction method-based CT image, the coeffcient of variation and contrast-to-noise ratio values in the lung nodule region showed similar values to those of the IR method. In addition, it was demonstrated that the point spread function value that can evaluate sharpness can be improved by using the FNLM filter. In conclusion, the results of this study are expected to maximize the image quality and reduce the dose by fusing the CT image reconstructed by the FBP method and the FNLM filter with excellent characteristics.

Jina Shim
Department of Diagnostic Radiology, Severance Hospital, 50-1, Yonsei-ro, Seodaemun-gu, Seoul, Republic of Korea

Myonggeun Yoon*
Department of Bioengineering, Korea University, 145, Anam-ro, Seongbuk-gu, Seoul, Republic of Korea

Youngjin Lee*
Department of Radiological Science, Gachon University, 191, Hambakmoe-ro, Yeonsu-gu, Incheon, Republic of Korea 
Study of a plastic s…
Study of a plastic scintillating plate‑based quality assu…
Introduction: The purpose of this study was to evaluate a plastic scintillating plate‑based beam monitoring system to perform quality assurance (QA) measurements in pencil beam scanning proton beam. 
Methods: Single spots and scanned fields were measured with the high‑resolution dosimetry system, consisting of a plastic scintillation plate coupled to a camera in a dark box at the isocenter. The measurements were taken at 110–190 MeV beam energies with 30° gantry angle intervals at each energy. Spot positions were determined using the plastic scintillating plate‑based dosimetry system at the isocenter for 70–230 MeV beam energies with 30° gantry angle intervals. The effect of gantry angle on dose distribution was also assessed by determining the scanning pattern for daily QA and 25 fields treated with intensity‑modulated proton therapy. 
Results: Spot size, field flatness, and field symmetry of plastic scintillating plate‑based dosimetry system were consistent with EBT3 at all investigated energies and angles. In all investigated energies and angles, the spot size measured was ±10% of the average size of each energy, the spot position measured was within ±2 mm, field flatness was within ±2%, and field symmetry was within ±1%. The mean gamma passing rates with the 3%/3 mm gamma criterion of the scanning pattern and 25 fields were 99.2% and 99.8%, respectively. 
Conclusions: This system can be effective for QA determinations of spot size, spot position, field flatness, and field symmetry over 360° of gantry rotation in a time‑ and cost‑effective manner, with spatial resolution comparable to that of EBT3 film.

Jaehyeon Seo, Myonggeun Yoon*
Department of Bio-Convergence Engineering, Korea University

Kwangzoo Chung, Youngyih Han, Sung Hwan Ahn*
Department of Radiation Oncology, Samsung Medical Center

Youngyih Han
Department of Radiation Oncology, Samsung Medical Center, Sungkyunkwan University School of Medicine, Seoul

Seonghoon Jeong
Department of Proton Therapy Center, National Cancer Center, Goyang

Yunhui Jo
Institute of Global Health Technology (IGHT), Korea University

Geon Oh, Yongha Gi, Heehun Sung
Department of BioMedical Engineering, Korea University, Seoul, Republic of Korea

Myonggeun Yoon*