Purpose: This study was
aimed to evaluate the utility based on imaging quality of the fast non-local
means (FNLM) filter in diagnosing lung nodules in pediatric chest computed
tomography (CT).
Methods:
We retrospectively reviewed the chest CT reconstructed with both filtered back
projection (FBP) and iterative reconstruction (IR) in pediatric patients with
metastatic lung nodules. After applying FNLM filter with six h values (0.0001,
0.001, 0.01, 0.1, 1, and 10) to the FBP images, eight sets of images including
FBP, IR, and FNLM were analyzed. The image quality of the lung nodules was
evaluated objectively for coefficient of variation (COV), contrast to noise
ratio (CNR), and point spread function (PSF), and subjectively for noise,
sharpness, artifacts, and diagnostic acceptability.
Results:
The COV was lowest in IR images and decreased according to increasing h values
and highest with FBP images (P < 0.001). The CNR was highest with IR images,
increased according to increasing h values and lowest with FBP images (P <
0.001). The PSF was lower only in FNLM filter with h value of 0.0001 or 0.001
than in IR images (P < 0.001). In subjective analysis, only images of FNLM
filter with h value of 0.0001 or 0.001 rarely showed unacceptable quality and
had comparable results with IR images. There were less artifacts in FNLM images
with h value of 0.0001 compared with IR images (p < 0.001).
Conclusion:
FNLM filter with h values of 0.0001 allows comparable image quality with less
artifacts compared
Jina Shim, Myonggeun Yoon*
Department of Bio-Convergence Engineering, Korea University, Seoul, Republic of Korea
Department of Diagnostic Radiology, Severance Hospital, Seoul, Republic of Korea
Mi-Jung Lee*
Department of Radiology and Research Institute of Radiological Science, Severance Children’s Hospital, Yonsei University, College of Medicine, Seoul, Republic of Korea
Youngjin Lee
Department of Radiological Science, Gachon
University, Incheon, Republic of Korea
Physica
Medica 81 (2021) 52-89